Constant Downstream Occlusion
RE: Infusion Pump Report Post
What line set are you using. Check occlusion sensor calibration low high and mechanical. Sounds simple.
Report Post
Dumb question however have you replaced the battery to see if that is causing your problem.
Report Post
ChemoMan, I literally covered all aspects of the occlusion detection system and I'm throwing you a dumb question? GTFOH, ChemoBoy
Report Post
The pump doesn't even attempt any infusion at all. It simply detects downstream occlusion. Yet, it will prime just fine...After replacing the sensor and completed calibrations, I'm here seeing if anyone has been through this before. Thanks.
Report Post
There are certain things the processor looks at during a reguar run and things it doesn't look at prime. The motor drive signal is different during prime. I think you have a processor problem or a motor problem. You might want to send it in to B-Braun. If you have been trained by B-Braun there techsupport is pretty good, otherwise no way are you getting help from them other than a RMA.
Matt D.
Report Post
If your battery doesn't hold correct voltage you can get all sort of problems also the door has a magnet in there make sure it's in tact
Report Post
If your battery doesn't hold correct voltage you can get all sort of problems also the door has a magnet in there make sure it's in tact