Sensor

Hi! I have this pump and it keeps giving me the distal air error I already modified the pot for the distal sensor but it keeps popping up the alarm, any advice? Thank you

   RE: Infusion Pump Report Post
Thu Dec 30 2021Reply from JPRV29
avatar placemark

In case anyone need it, you can check the pot called DA on the card, move it, I used 100 ohms and it worked, I just double check it with distal air on purpose and it works good. :)


Report Post
Thu Dec 30 2021Reply from JPRV29
avatar placemark

In case anyone need it, you can check the pot called DA on the card, move it, I used 80 ohms and it worked, I just double check it with distal air on purpose and it works good. :)


Report Post
Mon Jan 03 2022Reply from MedWrench Crystalyn
avatar placemark

What is the name of the manufacturer and model of the equipment you are working on? 


Report Post
Mon Jan 03 2022Reply from JPRV29
avatar placemark

It's a Hospira Plum A+ pump


Report Post
Wed Nov 23 2022Reply from Dave Lester
avatar placemark

If anyone needs it, you can verify the DA pot on the card by moving it. I used 80 ohms, and it worked. I then purposefully double checked it with distal air, and it functions properly.