Mon Mar 21 2022

need to calibrate a number of modules hosting on a M30 anesthesia monitor. Can I access through service mode on monitor, what is the service password?

need to calibrate co2 modules on MP30. What is service password?

   RE: Modules Report Post
Tue Apr 26 2022Reply from Chris S
Chris S

The default service code is 1345.   


Official MedWrench Guru
Report Post
Tue Apr 26 2022Reply from Ray Brown, CBET
avatar placemark

Yes. Do you have any of the Philips service manuals? Because the M3015A is referenced in several of the manuals, and the password is also mentioned. But a full PM does require some equipment. You need a low flow Thrope tube to verify 50 mL/min of flow into the module, 5% and 10% cal gas, and the appropriate tubings.

And the service password should be in the manuals. I can say it's a 4 digit password but I think they'll growl at me if I come right out and say it. :-)  


Report Post
Wed Apr 27 2022Reply from Atari
Atari

Ray Brown, CBET is right, instruments are needed for calibration. A password will not solve the problem.


Report Post