I get the message Flow obstruction.
Hoe can I fix it?
Fri Apr 01 2022
Troubleshooting
Fri Apr 01 2022Reply from Alberto Bea
Fri Apr 01 2022Reply from Chris S
Verify that the sample does not contain clots or aggregates. If no aggregates are visible, repeat once to rule out a defective PFA test cartridge.
• A sample that results in repeated flow obstruction may have resulted from problems during blood collection. Recollect blood sample and repeat test.
• If problems persist, contact Siemens Service.