Electrical Test Fails Code #117

Can anyone help to solve the Electrical Test Fails Code #117 problem on our CS300? Thank you in advance

   RE: Intra-Aortic Balloon Pump (IABP) Report Post
Tue Apr 05 2022Reply from Bill McSharry
Bill McSharry

This error code stands for:

117 Front End A/D reference failure at power up


Report Post
Tue Apr 05 2022Reply from Mark2022
avatar placemark

Ok, thank you very much 

how can i fix that? What need to be done? 


Report Post
Wed Apr 06 2022Reply from Denova Medical
Denova Medical

The problem is in the front end PCB or Main PCB. The front end PCB is a common failure on these devices. I would open the unit and check/wiggle/push all connections on Front End PCB and Main PCB, make sure chips are pushed into sockets and check it again. 

If failes, then check front end pcb reference voltage. From service manual below.

Reference Voltage Check
1. Refer to the calibration test point locations in FIGURE 4-15 on page 4-29. Connect a
DVM between TP22 and TP33 (GND_A) on the Front End board.
2. Adjust R253 on the Front End board for 5V ±0.1V.

If that does not work, swap front end PCB to get problem down to board level. Front end PCB is easy to replace, but you will need to do some calibrations after. At least swap it out and see if error goes away.  Denova Medical has front end PCB for sale.


Photios Peter Dalamagas Denova Medical Inc (704)-233-7574