Just changed the head. Now giving error 15.2.Engenier says that I need to change CPU I/O board and calport detector with cable. Plz help. Already spent a lot
Fri Apr 08 2022
High transmission error
Fri Apr 08 2022Reply from Batya
Sat Apr 09 2022Reply from Shazia Zubair
Sat Apr 09 2022Reply from Shazia Zubair
Sat Apr 09 2022Reply from Shazia Zubair
Sun Apr 10 2022Reply from bekzclz11
The transmission error is the variation of the deflection in the path of contact between pinion and wheel under load. It is calculated under static conditions for several meshing positions (n ⇒ 0).
Report Post
Sun Apr 10 2022Reply from Terry Hubbard
do a head and calport calibration. your technician should be able to do this, if he cannot get another Techniciam.
Terry
Report Post
Sun Apr 10 2022Reply from Terry Hubbard
do a head and calport calibration. your technician should be able to do this, if he cannot get another Techniciam.
Terry
Report Post
Wed May 25 2022Reply from Laserman
if the laser head was changed by someone and or the laser rod was replaced, the rotation of the rod needs to be addressed.
This is most likely the cause
Report Post