Sat May 07 2022
all parameters does not display after changing battery.
A few days ago sram error is on the screen. an all parameters does not display. Today I change the new sram battery and the new lead acid battery. And i did factory reset after reading official service manual. But still the all parameters does not display on the screen. What should I do? When the power on, self test is passed normay. But after self test, the all parameters does not dislpay. I think that this is the software setting problem, not a mechanical defect. Is there anyone who solved this same problem? Thanks lot. have a good day!
Tue Jun 14 2022Reply from BrentS
BrentS
1. Write down and/or take pictures of all parameter, unit settings before starting. Monitor will reset to basic factory settings and will need to be updated back to what they should be after SRAM battery is replaced
2. Install new SRAM battery.
3. I perform a factory reset, but is not always required.
4. Set monitor frame type.
5. Set monitor paramaters. Save and restart monitor.
6. Set all onscreen display paramaters, both analog and digital and save to mode type that is set for monitor startup.
7. Restart monitor to verify settings.
Fri Jul 15 2022Reply from BMETDoug
Thank you Brent!
Changing monitor frame type and restarting fixed my problem!