Grey display upon boot up
RE: System Report Post
Does it complete the first 5 steps fo the boot phase? Do you hear the test sound? If so, you probably have a display issue.
- Check cable connection is secure at both ends of the display flex cable.
- Check cable connection is secure at both ends of the backlight inverter cable.
- Replace complete Front Bezel Assembly
If it's not the display I would look at the API or main board.
Boot Phase Boot Phase Event
1 Switch the monitor on using the On/Off switch. (Green power light is lit as soon as the device is connected to AC mains.)
2 The green AC Power LED blinks for about during the boot phase.
3 You hear a ‘pop’ from the loudspeaker.
4 Green AC Power LED stops blinking, and remains lit.
5 Boot Screen with the Philips Logo appears on the display. Test Sound is issued. .
6 Boot Screen with the Philips Logo disappears. Fixed screen elements (for example smart keys, alarm fields) appear on the screen.
7 First measurement information appears on the screen, touchscreen is functional.