Sorvall - Legend T/RT Plus, Fisher Scientific accuSpin 3R
RE: Centrifuge Report Post
If the front panel is going through a reset simular to what it does on switch on then the reset is either being caused by a poor connection between the display and the main pcb or the rotor sensor is shorting out on the motor (the wires rub against the motor and eventually short out dropping the 24v
Report Post
We set this repair aside for a while, but I am on the case now. There is only one fan in this unit, and it is working well. When I turn it on it innitially seemed to work, then went into a reboot loop. There was a capacitor whine from the compressor start cap. I will replace that, but it does not seem like that should cause a reboot loop. I also checked the speed sensor wiring and it seems fine. The only other thing I can think of would be the brake resistor or thermal sensor. Any ideas?