Lid error

Hello, I have a Megafuge 1.0 that opens and closes perfectly when it is on standby. When you start the cycle, "LID" appears on the screen and remains blocked. The closure system is tested and correct. Any suggestion? Thanks.

   RE: Centrifuge Report Post
Tue Jun 14 2022Reply from Chris S
Chris S

Check the lid microswitch.  Take a resistance reading on it to make sure it is functioning and physically inspect it to make sure the arm isn't broken or bent.

You can also check page 2-3 of the Service Manual.


Official MedWrench Guru
Report Post
Wed Jun 15 2022Reply from celarrieta_5854
avatar placemark

The switches are ok, in fact opens and closes correctly in stand by, the problem is when you want to run a cycle. I will check the service manual instructions about the termal switch of the motor. Thank you!


Report Post
Wed Jun 15 2022Reply from celarrieta_5854
avatar placemark

The lid switches are ok, the motor over temperature fuse is ok too. I dont know to do.


Report Post
Wed Jul 13 2022Reply from Emily Vitkovitsky
Emily Vitkovitsky