Thu Jun 16 2022

Trouble shooting

I found this error 5004 and idont know what might be a possible causes Any one can help please??

   RE: Mobile C-Arm Report Post
Thu Jun 16 2022Reply from LISTER
LISTER

F5004 Gen. Uz ist out of range
Possible trouble: The unregulated intermediate circuit voltage is outside
+200V to +320V.
Possible help: Check voltage Uz on board D1. 
Measuring divider: 50 V corresponds to 1 V line voltage
wrongly programmed on transformer T1.
Check fuse D2/F1 and D2/F2.
D1 defective.
D2 defective.


Report Post