Might somebody be able to point me at instructions on how to replace the collimator bulb in a Fuji FCR GO machine?
Sun Jun 19 2022
Collimator bulb replacement?
Tue Jun 21 2022Reply from Chris S
1. Remove the caps 1 for the collimator knobs on the rear cover side. Remove the collimator knobs after loosening the M6 nuts.
2. Loosen the setscrew for the acrylic cross plate.
3. Detach the rear cover by removing the setscrews .
4. Detach the heat shield plate by removing the setscrews .
5. Detach the lamp cover by removing the setscrews .
6. Pull the lamp out of the socket.
7. Insert a new lamp Ë10 into the socket. The vinyl protection cover should be put on the lamp Ë10 during insertion so that you do not touch the lamp directly.
8. Assemble the parts by following the above steps in reverse order