Mon Jun 20 2022
DiagCode: 218.84PY.4BE3.37
Hello can you help me with this error DiagCode: 218.84PY.4BE3.37 softwarebuildversion: 5.1.0.206
Mon Jun 20 2022Reply from UL Service
Mon Jun 20 2022Reply from marcoscastro
Tue Jun 21 2022Reply from UL Service
1223 - control panel communication error.
Refer to service bulletin INT-060. Check these cables to the control panel: UMB -To- SPD cable, CP Segment 1 cable, and the CP Segment 2 cable. On each cable verify that the green and white wires (USB D+/D-) are tightly twisted together (four twist per inch or four twist per 2.54 CM) between the connector and the cable insulation. Verify each of the connector pins are fully engaged with the locking tab. Verify none of the control panel cable pins are bent or recessed. Replace cables as necessary. If the problem persists, replace either the control panel or the CPM board inside of the control panel.
Wed Jun 22 2022Reply from mike.ross.ultrasound@gmail.com
Hi, 218 Error:
AnnotSP Unanticipated Error
Description
The Annotation Signal Path has reported an error for which there is no
recovery. Scanning has stopped.
Corrective Action
This indicates a software error. Replacing hardware, or the current revision
software is unlikely to solve the problem.