Hello, I have this Hermle - Z206-A centrifuge that presents me with error 60, which is "Undervoltage in the intermediate circuit"
Does anyone know how it can be solved?
Tue Jul 05 2022
ERROR 60 Undervoltage in the intermediate circuit
Tue Jul 05 2022Reply from MapleTreeBiomed
The service manual lists the error with no actions to take. I just submitted the manual to be added to the documents section. It says the control board is also the power board. I'd check to make sure the power transformer switch is set to your local voltage 120 vs 240. After that it's likely a bad control/power board which is part number 914.010. I doubt you'll have any luck finding it though.
Good luck!