Error code 2202/243
RE: R/F Systems Report Post
243_2022_FLC-STU / XDR
"An STU-HW error occurred (12V PC power supply not ok)."
"IC an internal XD error occured"
Explanation:
An internal XD error occured.
Action:
Please reset the detector and restart the system
"243" relates to the Table/Fluoro Detector or Portable Detector or fixed Wall Stand detector
"STU" is the detector interface on the FLC system computer.
Report Post
I can't find an error code 2202 for location 32 in my code lists. The number 32 indicates a Tabletop Movement Control issue. So this could be Longitudinal or transverse directions. The transverse motor and feedback pot at located at the foot of the table. The longitudinal is located inside the table body, and it's comprised of the motor with a absolute encoder at the foot end, and a relative encoder at the head end.
Is 2202 the correct error number?
Report Post
Specifically, I can't find an error code 2202 associated with location 32. Is this an Agile?
Report Post