Mon Aug 08 2022

Aiming beam

Trying to repair my co2re. The aiming beam is not centered properly. I've got the thing opened up. Any advice on aligning the aiming beam properly? I'm having a tough time... Josh@thelaseragent.com

Equipment: Syneron-Candela - CO2RE


   RE: Aesthetic Laser Report Post
Mon Aug 08 2022Reply from Wayne-UML
Wayne-UML

First question: Are the aiming and treatment beams coaxial?

Second question: If they are, does the aiming beam stay normal when the arm is moved around? ("Normal" meaning not diffracted.)


Wayne Delgado
Senior Laser Engineer
www.umlasers.com
Report Post
Mon Aug 08 2022Reply from Laser J
avatar placemark

They are not. The laser beam is aligned, while the aiming beam is not. 


Report Post
Mon Aug 08 2022Reply from Wayne-UML
Wayne-UML

With the arm off, perform a near/far alignment of the aiming beam conforming it to the treatment beam, and you'll be good to go.


Wayne Delgado
Senior Laser Engineer
www.umlasers.com
Report Post
Fri Jun 16 2023Reply from CEST
avatar placemark

I am having the same issue, however when my aiming beam is moved around the beam does not stay "normal".



Report Post