Sat Sep 24 2022
Nellcore 595 err 6
Nellcore pulse oximeter showing err 6 , once I attach spo2 sensor
Equipment: Nellcor - OxiMax N-595
   RE: Pulse Oximeter Report Post
Tue Sep 27 2022Reply from T-Lobato
T-Lobato
I looked through the manual for a bit.  There are two things I would check:
-It appears there is a voltage switch on the back of the unit.  Make sure that is set according to your wall voltage.
-Check the critical battery.  It's internal.
Just me.
T.
TJ Lobato | MedWrench Guru
Tue Sep 27 2022Reply from Chris S
Chris S
6 SpO2 front end reports calibration (offset) failure
Service Manual is locate HERE.
I would try a new cable and sensor, then look at the UI PCB.