Error 57

What is this error mean!?

Equipment: Cynosure - Palomar Starlux 300


   RE: Aesthetic Laser Report Post
Tue Sep 27 2022Reply from Chris S
Chris S

ThermoController temperature sensor disconnected.

1. Ensure that the system has adequate space for air flow. Make sure that there are no air restrictions or blocked vents.

2. Let the system cool down.

3. Restart the system and let it run without emitting light for 2 minutes.

4. Call Palomar service.

1. If temperature is unstable or operating abnormally, replace power supply module.

 

Service Manual is available HERE.