How to empty vapor 2000

Equipment: Draeger - Vapor 2000/ D Vapor


   RE: Vaporizer, Gas Machine Report Post
Tue Nov 01 2022Reply from Chris S
Chris S

I will upload the operator's manual, but here are the instructions:

 

Vapor with keyed filling system If the Vapor is connected to an anesthesia delivery system:

1 Control dial must be engaged at »0«. If the Vapor is not connected to an anesthesia delivery system:

2 The control dial should remain engaged at »T«.

3 Hold an anesthetic agent-specific bottle for the appropriate agent below the drainage outlet at the bottom of the filling device

4 Set control dial at »T«, swing lever out. The sealing block should slide forward.

5 Using a 2.5 mm Allen key, open the drainage valve by turning it one or two times counterclockwise

? Drain anesthetic agent until the viewing glass is empty and no more anesthetic agent flows into the bottle. If the anesthetic agent in the wick also needs to be removed, see "Flushing the Vapor", page 53.