QC fails

I’ve changed the stepper motor and the motor control board then when I performed the QC it gave me black image in the beginning then the phantom appears shifted down and shrinked, please advise

   RE: Bone Densitometer Report Post
Thu Dec 01 2022Reply from DEXA Solutions
avatar placemark

There could be a number of issues.  Please contact our customer support so we can get more information to help diagnose.  info@dexasolutions.com


Report Post
Thu Dec 01 2022Reply from denis goncalves
avatar placemark

WhatsApp: +55 11 945400157

Denis Gonãlves


Report Post
Thu Dec 01 2022Reply from Jodiflo
Jodiflo

Where table calibrations performed after parts replament? This needs to

be done by a FE.


Jodi Andrews - Official MedWrench Guru 
Report Post
Thu Dec 01 2022Reply from denis goncalves
avatar placemark

call me on whatsapp, you need to calibrate the steps, there is anydesk

+55 11 945400157

Denis Goncalves

DK System Medical


Report Post
Sat Dec 03 2022Reply from Hossam Elfarghaly
avatar placemark

Dear Jodi,

All table calibrations performed after parts replament and there was a drift in the laser and we have corrected it then did the detector flattening successfully but still the QC fails as the spine of the phantom is minimized with large dark area in the end

 


Report Post
Wed Dec 07 2022Reply from Emily Vitkovitsky
Emily Vitkovitsky

If you are in need of immediate assistance please feel free to reach out directly to our sponsors. They are happy to help.

MERCO Biomedical

DEXA Solutions

Just be sure to tell them you were referred by MedWrench.

 


Report Post