Equipment: Mortara - ELI 230
not locking in settings
You're following this process?
Setting Date and Time
1. Select MORE from real-time ECG view.
2. Select CONFIG.
3. Select 3 Set Time/Date.
4. The preprogrammed date and time is displayed. To make changes to the date and time values (using a 24- hour clock) in the same format as displayed, use ? and ? to move back and forth through each row.
5. Use ? to select the year, month, day, hour, or minute.
6. Use +10 or -10 along with +1 or -1 to advance year, month, day, hour, or minute.
7. Select DONE to save changes.
8. Select DONE to return to the Configuration menu and DONE again to return to real-time ECG view
I did not see a reference to an internal battery but I would look on the motherboard for that and check it's voltage.
T.
TJ Lobato | MedWrench Guru
There is no internal battery on the main pcb. Its prettry simple internally. appears to get its power for the rechargable lead acid battery. That has been changed a few months ago. tested the battery this morning and it is fine. Load tested and I have an awesome battery tester.
Firmware versions are there newer version currently on 1.21.00
I am thinking is a pcb issue.
Everything else is working though? I would call Mortara (Hillrom) before I replaced the board. Might save you some bucks.