I have a table that when leveled back to neutral still has a slight tilt.
Equipment: STERIS - AMSCO 3085 SP
Equipment: STERIS - AMSCO 3085 SP
Either the table is mechanically loose or a limit switch needs to be repositioned. Check to make sure the table wobbles less them 1/4". It needs to move a little. If its not mechanically loose then the limit switch related to the tilt needs to be adjusted. The limit switches are a rocker arm with a roller. You may need to bend the arm a little and make sure the switch is mounted tightly.