I have a GE Panda warmer with a System Failure 09 - AA 10 xx xx. Manual suggests it's a power issue and to replace main board (I know...).
Replaced main board, all seemed well, failure returned.
Replaced power board, reset the heat engine, worked for the rest of a day, and failure returned.
Measured element resistance, it's good.
Measured both thermostats, they're good.
Collected error log through PuTTY and found this little nugget:
PWRM_3.0.0.29,5,36.5,100,M,0,0,0,-1,00,12
* SYSTEM FAILURE Ticks: 3207
* System Failure 09- Ticks: 3208
PWRM_3.0.0.29,5,36.5,100,M,0,0,0,-1,00,12
Any suggestions on where to go next? The output appears to resolve based on the subsequent output via serial connection, however, the screen remains hung up at the System Failure screen.
Equipment: GE Healthcare - Panda iRES