Need a solution

Have a steris 3080 rl/sp ot table there is issue of glitches when we doing up down moment its not working smoothly Please provide solution

Equipment: STERIS - AMSCO 3080 RL


   RE: Surgical Table Report Post
Mon Jan 09 2023Reply from Stanislav_
Stanislav_

first check the status of the relay in the control unit. second, check the lift drive or lubricate the shaft on the drive.


Report Post
Mon Jan 09 2023Reply from Chris-CBET
avatar placemark

do you smell any smoke from the base(bad pump), are you hearing any loud gurgling sounds(low on oil). is the issue you are seeing only with base up/down function or are you seeing it with other functions(foot, back, tilt, trendelenburg, etc)


Report Post
Tue Jan 10 2023Reply from Sankd
avatar placemark

Issue only base up down and this happened after changing oil shield because there is leakage issue


Report Post