Mon Jan 09 2023
Changing the infared rod
I have a babytherm 8010 it has 300 watt heating elements , but the available is 200 watt , is there any problem if i used these 2 elements?
Equipment: Draeger - Babytherm 8004/8010
   RE: Infant Incubators & Warmers Report Post
Mon Jan 09 2023Reply from CMB7
CMB7
From what I read on page 33  of: https://www.medwrench.com/documents/view/11605/draeger-babytherm-8004-8010-drager8004-8010-service-manual-pdf
Power to the heating elements  can be controlled by the temperature of the patient so that using 200 watts heating elements should not present a problem.
Regards, Michel.
Servicios Tecnicos CMB7
644 Merkley Drive,
Ottawa, Canada
Mobile: 6138185321
Registered Respiratory Therapist (Canada #72-659)
Tue Jan 10 2023Reply from Bishoy
Dear Michel, thanks for your reply
The engineer said that it gives error 35 , but It was un acceptable where the checking for the infrarein lements in error 71 , also according to the service manual the feedback signal controlled by triac and optocoppler   may be the signal is weak as the resistance of the 200 w is higher? And this makes conflict in the device ?
Thanks again