Equipment: Draeger - Babytherm 8004/8010
Changing the infared rod
RE: Infant Incubators & Warmers Report Post

From what I read on page 33 of: https://www.medwrench.com/documents/view/11605/draeger-babytherm-8004-8010-drager8004-8010-service-manual-pdf
Power to the heating elements can be controlled by the temperature of the patient so that using 200 watts heating elements should not present a problem.
Regards, Michel.
C. Michel Bessette, RRT (Canada #72-659)
Servicios Tecnicos CMB7
644 Merkley Drive,
Ottawa, Canada
serviciostecnicoscmb7@gmail.com
Mobile: 6138185321
Registered Respiratory Therapist (Canada #72-659)
Bio-Medical Technological Services: Compressed Gas Systems Specialist
Anaesthesia & Hyperbaric Chamber Technologist
Services in: English, Français & Español
Bio-Medical Technological Services: Compressed Gas Systems Specialist Hyperbaric Chamber Technologist Services in: English, Français & Español
Report Post

Dear Michel, thanks for your reply
The engineer said that it gives error 35 , but It was un acceptable where the checking for the infrarein lements in error 71 , also according to the service manual the feedback signal controlled by triac and optocoppler may be the signal is weak as the resistance of the 200 w is higher? And this makes conflict in the device ?
Thanks again