Changing ETC02 setting from % to mmhg

I need to change the ETC02 settings from % to mmhg .

Equipment: Datex Ohmedal Cardio Cap 5


   RE: All Categories Report Post
Wed Jan 11 2023Reply from MapleTreeBiomed
MapleTreeBiomed

Hey Harryb181,

Try after letting the monitor boot up fully.

Find Monitor Setup –> Install/Service (password 16-4-34) – Installation – Units – Parameters. If it's an option it should be there.

Otherwise, maybe make a cheat sheet? Find the ambient atmospheric pressure at your facility in mmhg and multiply it by the displayed percentage? So 760mmhg (pressure at sea level) * 5%= 38mmhg

Good Luck!


Report Post
Wed Jan 11 2023Reply from Chris S
Chris S

It should be as follows:

  1. Press the ComWheel and select Monitor Setup from the main menu
  2. Select Install/Service and enter 16 - 4 - 34
  3. Select Installation
  4. Select Units
  5. Select Parameters to review and/or change the units for a specific parameter.
  6.  Save changes in the Save Modes Menu using password 13 - 20 - 31

Official MedWrench Guru
Report Post
Wed Jan 11 2023Reply from Chris S
Chris S
  1. duplicate

Official MedWrench Guru
Report Post