I need to change the ETC02 settings from % to mmhg .
Equipment: Datex Ohmedal Cardio Cap 5
Equipment: Datex Ohmedal Cardio Cap 5
Hey Harryb181,
Try after letting the monitor boot up fully.
Find Monitor Setup –> Install/Service (password 16-4-34) – Installation – Units – Parameters. If it's an option it should be there.
Otherwise, maybe make a cheat sheet? Find the ambient atmospheric pressure at your facility in mmhg and multiply it by the displayed percentage? So 760mmhg (pressure at sea level) * 5%= 38mmhg
Good Luck!