Please help to solve the problem, ERR 0033
Equipment: Siemens - MULTIX COMPACT
Equipment: Siemens - MULTIX COMPACT
Possible Causes of Error:
• Temperature sensor has responded.
• High voltage transformer defective
Corrective Measures:
• Check temperatue of high voltage transformer
- If hot: 1) Operating error (limit operation) or 2) inverter defective -> replace inverter
- If cold: The temperature sensor in the HVT is defective; unplug the 14 connector on
the D500 and measure Pins 5 and 8 with the DVM. The value must be 47.5 kOhm at a
temp. of 25 °C (please note: at temp. < 25°C, the value becomes > ; at temp. > 25°C,
the value becomes < !).