Equipment: CR Konica Minolta Regius 110 HQ
Konica Minolta Regius 110, illegal bar code
RE: All Categories Report Post
Hello Albertkhamraev,
Can you confirm the cassette you're trying to use with the system?
Koby Kenney
MedWrench Guru
Hello Koby, I have 4 different sizes of original Konica Minolta Regius cassettes for reguis 110 HQ. The digitizer barcode reads 8x10 size and Mammo cassettes. It does not read (shows illegal barcode) 10x12 and 14x17 sizes. I have been using mostly these 10x12 and 14x17 sizes for almost 2 years with no problem but couple of days ago it stopped accepting them. We have not changed anything on settings. Everything is same as before. We have thoroughly cleaned barcode reader and cassettes, it didn't help. Thank you for your assistance in this matter
We have found the solution for Illegal barcode issue. LLB and LBP sensors need to be cleaned thoroughly.