you have a Baxter - Flo - Gard 6201 error code 49, what does it indicate and what should I do ?
Equipment: Baxter - Flo - Gard 6201
Equipment: Baxter - Flo - Gard 6201
Per the Service Manual :
Malfunction in real time clock: Abnormal RTC data.
Turn pump off and back on. Check backup battery voltage in “Automatic Test Mode 1: Calibration Mode 1” on page 9-14. Reset all configuration option settings in “Modifying the Configuration Option Settings” on page 5-2
I would try replacing the main and backup batteriesand then resetting the options such as the date and time in the configuration menu. Last option would be the CPU board.