Error code

We are having an issue with one of our units. It is displaying the message CLV lamp ERR E103. Please check examination lamp. I have not seen this before. The device is currently in use. It does not seem to be hindering the department from using it. Thank you.

   RE: Light Source Report Post
Sat Feb 18 2023Reply from Stanislav_
Stanislav_
Hello! This message indicates that the lamp needs to be replaced. Or you can reset the number of hours. for this you need to enter the settings.

Report Post
Tue Feb 21 2023Reply from Michael78
avatar placemark
Thank you. I did replace the lamp. I may have forgotten to reset the hours. I will look in to that,

Report Post
Wed May 03 2023Reply from swami
avatar placemark

Hi Stanislav

I have error at olympus CV-290 error E300 .. please let me know what is this error means.


Meera

Report Post