Hello everybody! Does anyone know the reason for err08? Share who can scheme on this centrifuge. Thank you! serega.mez.88.ms@gmail.com

Equipment: Thermo Scientific - Heraeus Labofuge 400


   RE: All Categories Report Post
Fri Feb 24 2023Reply from Chris S
Chris S
Error Cause: Over-voltage of intermediate circuit. Possible Error Source: Defective leads or brake resistor, Driving and/or braking circuits. Corrective Procedure: Replace the main board, Check leads and brake resistor, replace defective parts.

Official MedWrench Guru
Report Post
Sat Feb 25 2023Reply from Serega.88.kz
avatar placemark
Thank for the help