Does anyone have the output testing procedure or the calibration procedure for a Surgistat II (SURGII-20).
I looked through the User’s Guide and the service manual and have not found what I am looking for.
The REM function needs to be calibrated, and I feel like the output verification in the Service manual should include output power verification using an ESU analyzer. Any tips would be apprectiated.
Wed Mar 01 2023
Output Testing and Calibration Procedure
Thu Mar 02 2023Reply from Chris-CBET
I have the 1980s service manual with what you're looking for, unfortunately it isn't a copy that can be cut n copied. Once medtronic took over they cut out the entire section. What's your email I send it to you, or look for it on Frankshospitalworkbench
Report Post
Fri Mar 03 2023Reply from MJ-925
Wed Mar 15 2023Reply from MJ-925
Mon Mar 20 2023Reply from Chris-CBET
hi Matt, I did, I sent it on March 3rd, you replied with a cc to Tanner to which he started asking me about someone named Juan from some other company i'm not affiliated with