Does anyone have the output testing procedure or the calibration procedure for a Surgistat II (SURGII-20).
I looked through the User’s Guide and the service manual and have not found what I am looking for.
The REM function needs to be calibrated, and I feel like the output verification in the Service manual should include output power verification using an ESU analyzer. Any tips would be apprectiated.
Wed Mar 01 2023
Surgistat II Output tests and calibration Procedure
Wed Mar 01 2023Reply from Emily Vitkovitsky
If you are in need of immediate assistance please feel free to reach out directly to our sponsors. They are happy to help.
SAKOMED: https://www.medwrench.com/service/companies/sakomed-275
Integris Equipment: https://www.medwrench.com/service/companies/integris-equipment-239
A & G Biomedical: https://www.medwrench.com/service/companies/a-g-biomedical-273
Just be sure to tell them you were referred by MedWrench.
Report Post
Thu Mar 02 2023Reply from Chris S
I normally verify the output using the Graphs starting on Page 8-8 of the user manual, or page 3-20 of the service manual. Verify outputs at a few different loads and powers.
REM function can be checked with an ESU analyzer or Decade box using the technical specifications listed in the service manual.
https://www.medwrench.com/documents/equipment/1124/valleylab-surgistat-ii
Official MedWrench Guru
Report Post