On pressing the power button the fan turns on but does not boot then after 1-2 minutes it switches off

Equipment: Philips cx50


   RE: Ultrasound Systems Report Post
Mon Mar 06 2023Reply from Emily Vitkovitsky
Emily Vitkovitsky
If you are in need of immediate assistance please feel free to reach out directly to our sponsors. They are happy to help. Encore : https://www.medwrench.com/service/companies/encore-technical-services-inc-276 Integris Equipment : https://www.medwrench.com/service/companies/integris-equipment-239 Just be sure to tell them you were referred by MedWrench

Report Post
Mon Mar 06 2023Reply from Katherine Mazza
avatar placemark
Could be the power supply is dead, or more likely the power regulator board is bad.

Report Post
Fri Mar 10 2023Reply from AyoCT
avatar placemark
The battery is bad and needs to be replaced. Thanks

Report Post
Fri Mar 10 2023Reply from WSawalha
avatar placemark
It does the same with the power supply connected