Error 4021

Hello, I got error 4021 from Siremobil Compact L with Memoskop CX, The text appear on monitors. The memoskop function CTRL+T, CTRL+U work with generator connected. I have checked the X10 'main umbilitcal cable' connection pins for damage/bent over, and none of the pins are pushed-in, a couple of pins are partly-in, this is normal. I have checked the cable wires between the X10 & Memoskop in case a wire is broken. I tried to Memoskop, Technical setup, menu "Diagnostics" - "Erase Harddisk" but doesn’t has “Diagnostics” menu, Please advice. Best Regards, Chao.

   RE: Mobile C-Arm Report Post
Sun Mar 26 2023Reply from outinleft
avatar placemark
The message "minimum mode" should show on the monitors for about 30-60 sec. after booting. Does it disappear after about 30-60 sec. as it should?

Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com
Report Post
Mon Mar 27 2023Reply from Chaovalit
avatar placemark
Mr.Wayne Horsman, Thank you for your kindly support. The message "minimum mode" doesn't show. It looks normal. My last try, I swapped the memoskop to another unit that the error follow to other unit (I try quickly cause I have 10 minute allowed from my customer to try.) Best Regards, Chao.

Report Post
Mon Mar 27 2023Reply from Emily Vitkovitsky
Emily Vitkovitsky
If you are in need of immediate assistance please feel free to reach out directly to our sponsors. They are happy to help. Agiliti: https://www.medwrench.com/service/companies/agiliti-295 Smart Choice Medical : https://www.medwrench.com/service/companies/smart-choice-medical-usa-llc-290 Just be sure to tell them you were referred by MedWrench

Report Post
Mon Mar 27 2023Reply from Chaovalit
avatar placemark
Thank you very much.

Report Post
Sat May 06 2023Reply from lorde
avatar placemark
fixed it according to your instructions thanks a lot.


buckshot roulette
Report Post
Mon May 08 2023Reply from outinleft
avatar placemark

Glad I could help


Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com
Report Post
Mon Aug 21 2023Reply from Ralph Frizzle
Ralph Frizzle

WE HAVE MANY SIEMENS COMPACT AND MEMODKOP PARTS IN STOCK


Ralph Frizzle
Radiology Engineer
Ralph@GermanElectronics.com

Report Post
Thu Aug 24 2023Reply from profescalante
avatar placemark

Saludos tengo roblemas con uno de estos equipos luego que en una operacion usaron un taladro y lleno de sangre la unidad de la camara, luego de ello alguien lo abrio y trato de limpiarlo al no poder ponerlo a funcionar nos llamaron pero al llegar y encender el equipo y al realizar una exposicion no generaba ninguna imagen, al nosotros abrirlo nos dimos cuenta que el microsuiche que la pantalla estaba suelto, le hicimos el mantenimiento colocamos el micro en su sitio calibramos acto seguido encendimos la unidad se hicieron todas las pruebas con el puente todo opera bien sube baja realiza todas las funciones entonces hicimos las exposiciones pero las imagenes son de mala calidad estoy pensando que la unidad sufrio ustedes que opinan en que me pueden colaborar a ver si restablecemos el servicio de este equipo muchas gracias.


Report Post