100 W Output limit for Force 2

This Force 2 (Dash number 20CEM) will not output any monopolar power higher than 100 Watts. The power settings can be adjusted to the maximum values for each mode on the display (300 - pure cut, 250 - Blend 1, 200 - Blend 2, 150 - Blend 3, 120 - Coag). But when the footpedal or handswitch is activated, the unit will reset the power setting to 100 W, then output power at that setting. The Bipolar mode does not seem to be affected by this. This seems like a setting that needs to be adjusted. Any input is appreciated.


   RE: Electrosurgical Unit (ESU) Report Post
Fri Apr 28 2023Reply from Aneserv Medical Inc.
Aneserv Medical Inc.

Did you try to just use the pencil without the footswitch attached? I have seen the conmed foot petal attached by accident to these units as they will fit perfect but the cut/coag pedals are backwards...should have been a recall on this years ago! Otherwise, you are probably looking at a bad Transister on the output board.


Report Post
Tue Oct 24 2023Reply from nbaum
avatar placemark

I have the exact same issue with the mono cut resetting to 100 watts every time its activated. Except on mine its only the "CUSA Accessory" outlet when I use a switched handpiece or footswitch that has the issue. The "handswitch" receptacle on the right does not have this issue with the handpiece or using the foot switch.


Did you ever find the culprit or a solution? Thanks!