Thu May 18 2023

F46 error

What is the F46 error on Lumenis Selecta II laser and how do I resolve it?  Thanks in advance for any suggestions!


   RE: Surgical Laser Report Post
Thu May 18 2023Reply from Laserdude
avatar placemark

f46 is a calibration error.

The high voltage needs to be checked.  If it is found to be below 600vdc, most attempts at recalibration will be fruitless in the near future.

The laser cavity may need to be rebuilt.


Report Post
Thu May 18 2023Reply from lillyeye
avatar placemark

Thank you for your reply.  I am not certain I know how to recalibrate laser.  Do you do this service or can you recommend someone?


Report Post
Mon May 22 2023Reply from Emily Vitkovitsky
Emily Vitkovitsky

Hey, I'm sorry to hear you're having equipment issues. Please fill out aREQUEST FORM and we will pass it along to a company to contact you directly. Our vendors can help service your equipment. We are excited to help!


Report Post