Mech error

Unit passes initialization and is able to run samples. Fails start up every time. There is a "Mech error: Motor Syringe gap".... Anyone encountered this and remedied? Any help is greatly appreciated.

Equipment: Abbott - CELL-DYN Emerald


   RE: Hematology Analyzer Report Post
Mon Jul 31 2023Reply from Paul Moyer
avatar placemark

You might need to clean and grease the white syringe pistons. Only use the grease supplied by Abbott.