Sat May 27 2023

fault 12.2

what is fault 12.2


Equipment: Syneron-Candela - GentleMax Pro


   RE: Aesthetic Laser Report Post
Tue May 30 2023Reply from MapleTreeBiomed
MapleTreeBiomed

Hi Janice,

The manual says fault 12.2 is a High Energy fault."Laser head energy of last treatment pulse was high." and suggests to recalibrate the laser system. If it persists contact Candela tech support.

Good luck!


Report Post
Thu Jun 15 2023Reply from Candela Guy
avatar placemark

Needs calibrating again by a service engineer  first. Also try and remember if it does it on Yag only or Alex,


Report Post