The equipment turns on and works normally until we inflate the cuff and then it generates an alarm and only 930 appears on the screen, what could be causing this?
Equipment: dinamap v100
The equipment turns on and works normally until we inflate the cuff and then it generates an alarm and only 930 appears on the screen, what could be causing this?
Equipment: dinamap v100
If you are in need of immediate assistance please feel free to reach out directly to our sponsors. They are happy to help.
Just be sure to tell them you were referred by MedWrench
Per the service manual:
Error Code: 930
Source: SPO2
Definition: No status from module for 30 ±10 sec. Fatal error reported by module
Can Be Silenced: No
Probable Source: Check cable and sensor; Parameter turned on - no hardware installed in unit