CLINICA DAVILA

The equipment turns on and works normally until we inflate the cuff and then it generates an alarm and only 930 appears on the screen, what could be causing this?

Equipment: dinamap v100


   RE: Vital Signs/Multi-Parameter Report Post
Wed May 31 2023Reply from Emily Vitkovitsky
Emily Vitkovitsky

If you are in need of immediate assistance please feel free to reach out directly to our sponsors. They are happy to help.

Sage Services Group

Pioneer Biomedical

USOC

 

Just be sure to tell them you were referred by MedWrench


Report Post
Wed May 31 2023Reply from AJ_Diaz
avatar placemark

Per the service manual:


Error Code: 930

Source: SPO2

Definition: No status from module for 30 ±10 sec. Fatal error reported by module

Can Be Silenced: No

Probable Source: Check cable and sensor; Parameter turned on - no hardware installed in unit


Report Post