I have replaced the Locking Assembly and the Main Board
Still getting E 53 after the lid is closed and locked into place
Anyone ever have this issue?
I have replaced the Locking Assembly and the Main Board
Still getting E 53 after the lid is closed and locked into place
Anyone ever have this issue?
If you are in need of immediate assistance please feel free to reach out directly to our sponsors. They are happy to help.
Just be sure to tell them you were referred by MedWrench
There are sensors in the plastic body next to the control panel that sense the magnet in the lid to sense closure. Check those sensors Maybe one is dislodged?
I have 2 of the 20R refrigerated units on my workbench. Google me if you want to chat.
Don Settergren 2663
Inside Outside Sales LLC
Fitchburg WI 53711
Or, one of the sensor wires is in the wrong socket on the mainboard. I just checked and it looks like there are three identical sockets on the mainboard for those two sensor connectors!