Fri Jun 02 2023

E 53 issue

I have replaced the Locking Assembly and the Main Board

Still getting E 53 after the lid is closed and locked into place

Anyone ever have this issue?


   RE: Centrifuge Report Post
Mon Jun 05 2023Reply from Emily Vitkovitsky
Emily Vitkovitsky

If you are in need of immediate assistance please feel free to reach out directly to our sponsors. They are happy to help.

Ozark Biomedical

Agiliti

 MERCO Biomedical

Just be sure to tell them you were referred by MedWrench


Report Post
Mon Jun 05 2023Reply from T-Lobato
T-Lobato

Can you test the control panel?  


TJ Lobato | MedWrench Guru
Report Post
Thu Jul 27 2023Reply from dsettergren_1657
avatar placemark

There are sensors in the plastic body next to the control panel that sense the magnet in the lid to sense closure. Check those sensors Maybe one is dislodged?


I have 2 of the 20R refrigerated units on my workbench. Google me if you want to chat.


Don Settergren 2663

Inside Outside Sales LLC

Fitchburg WI 53711


Report Post
Thu Jul 27 2023Reply from dsettergren_1657
avatar placemark

Or, one of the sensor wires is in the wrong socket on the mainboard. I just checked and it looks like there are three identical sockets on the mainboard for those two sensor connectors!


Report Post