Hello, I have a problem with a bed, the knee down, head down, foot down functions do not lower their respective functions. The bed only performs the elevation functions but does not do any lowering


   RE: Bed Report Post
Tue Jun 20 2023Reply from Chris-CBET
avatar placemark

have  you tried disconnecting the siderail cables. 


Report Post
Wed Jun 21 2023Reply from Emily Vitkovitsky
Emily Vitkovitsky

If you are in need of immediate assistance please feel free to reach out directly to our sponsors. They are happy to help.

Agiliti

Integris Equipment

MERCO Biomedical

 

Just be sure to tell them you were referred by MedWrench


Report Post
Fri Jun 23 2023Reply from marianomedina_6724
avatar placemark

Hi Chris, so far I have checked the voltages of the solenoid valves and the motor, in that everything is fine, I thoroughly cleaned the hydraulic manifold and replaced the hydraulic oil, verified that the cables that reach the solenoid valves are correct, however the bed continues to fail and does not do the positions well


Report Post
Fri Jun 23 2023Reply from Chris-CBET
avatar placemark

I'm wondering if you have a stuck button situation which is common on these beds. disconnecting the siderails will tell you if that is your issue or not