Equipment: BioSystems - BTS 330


   RE: Clinical Chemistry Analyzer Report Post
Wed Aug 30 2023Reply from sachidan
avatar placemark

Replace lamp , If absorbance out of limits on all filters , if not , absorbance out of limit for one particular filter change the filter , Most probably the this problem will arise for 340nm (trhe filter has become opaque). 

2. Clean the Flow cell , throughly,  The absorbance of  Flow cell with  distilled water , shall not exceed 0.3 in 340 nm