Error Code ARC1301 (10855) and parallel lines atrefact

Sir Please give me information showing error code ARC1301 (10855) 

This error showing ones time  after that i started Digitizer 

Then not show error 

But image showing parallel lines atrefact every times 

Can you give me feedback on 8859295297??


Equipment: Agfa CR 10X


   RE: Computed Radiography Report Post
Wed Jun 28 2023Reply from Emily Vitkovitsky
Emily Vitkovitsky

If you are in need of immediate assistance please feel free to reach out directly to our sponsors. They are happy to help.

MERCO Biomedical

Tri-Imaging Solutions

 

Just be sure to tell them you were referred by MedWrench


Report Post
Sat Jul 29 2023Reply from deesmithk
avatar placemark

Look at the image plates within the cassettes. I am not sure if the cr10 has a dust brush like the cr30, but look inside the front of the unit, if a brush is visible get it out and look for an opening on the side of the reader. Then push brush in and out several times and see if the lines move or go away. If there's no brush, then open the reader and start looking for any object or dust that might be in the path of the image plate.


Report Post
Mon Aug 07 2023Reply from deesmithk
avatar placemark

the error arc 1301 = barcode label was not read. 

1) try another cassette

2) protect the unit from sunlight and room light

3) dust brush is not in the park position

4) there is no barcode on the plate, or it is not readable

5) the PMT and light detector may be defective (replace)

6) Optic module is defective (replace)


Report Post