Generator Non-thermal Inhibit

Hello All, I'm getting a "generator non-thermal inhibit" message on my GE Optima 220 portable x-ray machine.  This usually clears after rebooting machine, but is coming back more frequently.  Any idea as to what may be the issue?  Thanks!

Equipment: GE Optima 220


   RE: Mobile Radiography Report Post
Wed Jul 12 2023Reply from Emily Vitkovitsky
Emily Vitkovitsky

If you are in need of immediate assistance please feel free to reach out directly to our sponsors. They are happy to help.

MERCO Biomedical

Tri-Imaging Solutions

 

Just be sure to tell them you were referred by MedWrench


Report Post
Thu Jul 13 2023Reply from jrreyes
jrreyes

Thats a generic error. Try running generator diagnostics


Report Post
Thu Jul 13 2023Reply from TheVillages
avatar placemark

Replaced the Aux Box to rotor cable and issued was resolved.  Thanks All.