Just wondered if anyone has an idea what will cause the code?
Equipment: Sigma Spectrum infusion pump
Just wondered if anyone has an idea what will cause the code?
Equipment: Sigma Spectrum infusion pump
You can try cleaning and calibrating the slide clamp. Calibration is in the biomed menu, code 8961. Make sure you are using a new authentic "Baxter Blue" slide clamp. The color can fade over time causing the calibration to be off.
I would try that if I could get passed the error screen to the options menu. It powers up to the error screen and I don't see a way past it. Any idea's?