Equipment: RF8000E
How to test

I will upload the instructions, but here is the majority of it:
The following equipment is required to perform the power output test on the Coblator II System.
- Electrosurgical Analyzer (Dynatech 454 or equivalent)
- Wand with Patient Cable or adapter cable P/N 33102
Step 1 – Set the Electrosurgical Analyzer to the following parameters: 1. Select Manual mode 2. Select Output generator test 3. Set internal load to 300 ohms
Step 2 - Set up the Coblator II System as stated in the User’s Manual. Connect the “Active” input from the ESU to the tip of the Wand and the “Dispersive” or “Return” to the shield of the Wand.
Step 3 - Choose the setpoint (1-9) for either ablation mode or coagulation mode and press the foot pedal. Verify the readings with the table on the following page.
Load = 300 Ohms Units of measure = RMS watts Ablation Mode Setpoint Range (+/- 20%)
1 22-32
2 32-48
3 50-76
4 70-106
5 96-144
6 122-184
7 153-231
8 192-288
9 220-330
Load = 300 Ohms Units of measure = RMS watts Coagulation Mode Setpoint Range (+/- 20%)
1 9-13
2 11-17
3 13-19
4 14-22
5 16-24
6 21-31
7 22-34
8 25-37
9 28-42