How to test RF8000E?

Equipment: RF8000E


   RE: All Categories Report Post
Tue Jul 25 2023Reply from Chris S
Chris S

I will upload the instructions, but here is the majority of it:


The following equipment is required to perform the power output test on the Coblator II System.

  1. Electrosurgical Analyzer (Dynatech 454 or equivalent)
  2.  Wand with Patient Cable or adapter cable P/N 33102


Step 1 – Set the Electrosurgical Analyzer to the following parameters: 1. Select Manual mode 2. Select Output generator test 3. Set internal load to 300 ohms 

Step 2 - Set up the Coblator II System as stated in the User’s Manual. Connect the “Active” input from the ESU to the tip of the Wand and the “Dispersive” or “Return” to the shield of the Wand. 

Step 3 - Choose the setpoint (1-9) for either ablation mode or coagulation mode and press the foot pedal. Verify the readings with the table on the following page.

Load = 300 Ohms Units of measure = RMS watts Ablation Mode Setpoint Range (+/- 20%) 

1 22-32  

2 32-48

3 50-76 

4 70-106 

 5 96-144 

6 122-184  

7 153-231  

8 192-288 

9 220-330 

 Load = 300 Ohms Units of measure = RMS watts Coagulation Mode Setpoint Range (+/- 20%) 

1 9-13 

2 11-17 

3 13-19 

4 14-22 

5 16-24 

6 21-31 

7 22-34 

8 25-37 

9 28-42