C05 error

Occasional C05 (unstable signal) when starting BP.  Issue is not related to cuff. Thank in advance.

Equipment: welch Allyn spot vital sign


   RE: All Categories Report Post
Thu Jul 27 2023Reply from Denova Medical
Denova Medical

Do an NIBP calibration with the Welch Allyn Service tool.