Does anybody know step by step way to calibrate an easy 11 plus thank you in advance.
Or better off send me a link to a video on how to do it
Equipment: Tuttnauer - EZ11 Plus
Sat Aug 05 2023Reply from MapleTreeBiomed
MapleTreeBiomed
Hey MarkBassi,
Tuttnauer really likes keeping that info and the needed tools away from people unless they have gone to the tauttnauer school. Sometimes we use the term "calibrate" interchangeably with annual maintenance steps. So just be clear. You are asking how to adjust temp and/or pressure offsets, or just to do the annual servicing?
Mon Aug 07 2023Reply from Ross
Did you check the Documents, here?
Mon Aug 07 2023Reply from Emily Vitkovitsky
Emily Vitkovitsky
You can find the manuals HERE
Thu Aug 17 2023Reply from MarkBassi
Sorry for the slow response, and yes I am looking to set the pressure and temperature offsets I have the calibration kit
And all the tech books from tutt.
Thu Aug 31 2023Reply from patientcrista
All the best
Thu Aug 31 2023Reply from bayside1
First you need to have a good # for both Temp and Pressure in the chamber. You do not need Tuttnauers test kit for this.
I used my Madgetech temp and pressure data loggers.
Once tou have those #"s you enter them in to the appropriate screen.
At that point you are telling the machine what the actual #'s in the chamber are versus what is read on the screen.
The program will adj. the numbers.