Rev F.3 , it will start to boot for about 4 seconds, green led comes on near fans then shuts down.
Equipment: Philips - iE33 Echocardiography
Rev F.3 , it will start to boot for about 4 seconds, green led comes on near fans then shuts down.
Equipment: Philips - iE33 Echocardiography
I would check voltages on power supply and adjust if needed. Since its only powering on for a couple sec it may be tricky to catch and adjust. but here is what you should get if you need a diagram provide you email and I can send it over.
+3.65 V = 3.70 - 3.75V
+5.35 V = 5.35 - 5.40V
-5.35 V = -5.35 - -5.40V
+12 V = 12.00 - 12.05V
+15 V = 15.10 - 15.15V
-15 V = 15.05 - 15.10V
+48 V = 47.90 - 48.10V
more info. I would think if the light in the back isn’t flashing when it shuts down, it could be a faulty power button or usb board behind the dvd drive.remove the connections there and power on from the signal distribution on the wall beside the umb/emb. If that doesn’t work or the light is flashing, it could be any one of the following.
Power supply
Burned pins on AFP
AFP to Power supply cable
NAIM or Channel Board